Filter measuring wrong results #240
Replies: 4 comments 9 replies
-
Hi,
I wouldn't worry too much about a 0.145 dB deviation from the expected value. Yes, it shouldn't be there and yes, the R&S measurement gets you a better result but its price point is also vastly different from the LibreVNA. The frequency range is also not ideal for the LibreVNA. It can take measurements there, but the dynamic range and accuracy will be better above 10 MHz. The -0.428 dB is a bit more interesting. My first guess is that you are running into an overload condition here. Do you have an "ADC overload" warning in the bottom right corner? If this warning is there either during the calibration or during the DUT measurement, it could easily throw off your measurement by half a dB (or even more). You need to reduce the power to get rid of that warning. One other thing I can think of is port mismatch on the LibreVNA which could result in some additional reflections on the cable. This will show up as valleys and peaks in the frequency response. Their locations change depending on the cable (or DUT) length. Inserting your DUT will change this and a peak may end up where a valley was before, resulting in a positive dB value. But in general, these reflections come into play above 3 GHz with the LibreVNA, so it is somewhat unlikely to be the problem here. |
Beta Was this translation helpful? Give feedback.
-
Hi, I'm worrying a lot about the 0.145 dB in insertion loss! I allways use -10 dBm, the 0 dBm was just a test where I've been wondering that every change in power with a new calibration is showing different results. This should not be! Of course I reboot the computer several times and also the VNA after Firmware change from 1.2.1 - 1.5.0. |
Beta Was this translation helpful? Give feedback.
-
Well, of course your filter will have no gain. What I meant is that I wouldn't worry about such a small deviation from the expected value (your filter has probably very little insertion loss). If you are really concerned with such precise measurements, the first thing I would do is create a valid calibration kit file. In your attached calibration, you are using ideal standard (the default in the GUI if you don't change that). This also means that your through is expected to be of length 0 and has no insertion loss at all. This in itself should not cause a positive dB value for S21, it just introduces additional uncertainty because your standards likely are not ideal (although they are probably fairly close to ideal at such low frequencies). One thing you could try to rule out or confirm my suspicion of port mismatch: attach attenuators (e.g. 3dB) at both ports of the VNA. Then repeat your calibration and measurement with these attenuators. If your measurements are now better, there could be some reflection from the VNA ports. One other thing you could do to sanity check the measurement is measuring a different cable than your through standard (preferably approximately as long as your DUT) and check if that one will also have positive dB values. |
Beta Was this translation helpful? Give feedback.
-
Hi, |
Beta Was this translation helpful? Give feedback.
-
Hi,
a very simple thing to measure a lowpass filter between 500 kHz and 8 Mhz.
I calibrate SOLT on both ports, with the through cable I read S21 0,000dB as it should be.
The through cable is 30 cm with SMA, the connection to the DUT is 2x 15 cm one side each solderd to the PCB.
So exakt the same cable just cutted into two peaces.
With the DUT connected I can read 0.145 dB for S21 / S12.
This can not be right! It must be negative, an attenuation must be -0.145 dB.
Of course I get a negative reading in the stop band.
I tried 0 dBm and after a new calibration I got -0.428 dB S21 / S12.
This is also wrong, this is too high.
Measured with a RS FSH4 S21 is 0.08 dB, same Short Open Load and cable.
1st why is the resulut of S21 positive at -10 dBm?
2nd why is the result of S21 different at 0 dBM and too high?
3rd why is the result changing why changing the power for the measurement?
The calibration is done on each change of power.
The DUT is Chebyshev Lowpass Filter 5th order, nothing new.
Just want to do some measurings for a documentation.
Tried with 1.2.1 and 1.5.0. with the matching firmware for each version.
Beta Was this translation helpful? Give feedback.